I've tried using very small sample sizes (10-15 mg) for TVT/FET but reproducibility in results is non-existant. I bumped up sample size to ~ 100 mg and have done multiple extractions on the same sample. Where I'd expect to see a regression curve, I see a horizontal line or even an increase. My GC run time is 30 minutes (I have no backflush capabilities), so I have the vial taken out of the incubation chamber. When the vial comes out, there is no visible liquid so, even with this somewhat large sample size, I'm postulating that I'm still volatilizing all of the material. If you're investigating a hodge podge of materials, what would you expect the results of MHE would be?
Here's what I'm using:
Combi-Pal with Static Headspace (2 mL syringe) and incubation chamber
Incubation: 100 °C for 15 minutes
Needle: 105 °C
Inlet: 250 °C
Column: 40 °C initial for 4 min, ramp 10°/min
Constant flow
Sample: Ethanol, toluene, xylenes, naphtha, random petroleum distillates
Prep: 20 mL Headspace vial, 50 mg ethylene glycol diethyl ether (ISTD), 50 mg sample
I don't have the mhe attachment, so I come over and vent the vial manually after every injection.
Eventually, I want to be able to perform analysis on paint samples as a modification of ASTM D6886. I'm tired of ruining columns and liners from extracted materials into methanol and it makes sense to test volatile compounds via headspace, but this thing is giving me a real headache. Any help would be greatly appreciated.
Thanks!