INJECTION is NEAT...
1. Column used is 10% OV-225 6 ft x 1/8 in x 2.1 mm
2. Only column available for second methodology is 10% OV-101 6 ft x 1/8 in x 2.1mm
I do not think this is going to work using that second column. I have been doing a lot of work on this and the EOEMA is polar and the OV-225 is polar. and the method is a gradient temp beginning at 80 C with a initial time 2.00 min and rate of 10 C/min with final T = 200C.
I do not think I can use this 10% OV-101. It is a non-polar column and the only mechanism for separation of this EOEMA is by boiling point. I do not think this will give me a clear picture of all the impurities in the sample and I need those to do % Purity determination.
Think I am going to have to request a column that is polar..Something like a Carbowax with PEG...Maybe even a standard polar packed column. I think anything else is going to require the method be changed. Have to avoid that at all cost. I see a lot of options..could even use the DB-WAX capillary column I just ordered but would have to change the method ESPECIALLY flow rate. Cannot use 30 cc/min Helium with a capillary column.
This is a s SPECIAL CASE for validation and I have to compare 2 methodologies. All operating parameters are kept constant have to implement a different column with different retention characteristics but keep all other parameters in the method constant.
Please definitely let me know your thoughts. I have spoken with Agilent and Supelco...So I really do not think this OV-101 will fly....