I've been looking at standard addition methods for a while now but have not had the time/courage to try one. However I have been forced into it due to problems beyond my control. As such I though I would post what I've done to see if I've done things right, or if I've made some rookie mistake that i can't see.
The samples are water soluble products that are testet for ethanol content (5000 ug/g limit or 0.5%). 1g is disolved in 10 mL distilled water (volumetric flask). I then pipetted 1000 uL into 3 sample vials. To the first I added 500 uL of distilled water. To the second I added 250uL of a standard that I made up that corresponded to 1250 ug/g ethanol and 250 uL distilled water. The third sample recieved 500 uL of the standard which corresponded to 2500 ug/g ethanol. I then injected the 3 samples. The R2 was 0.996 and the final concentration (thanks Excel) was 4260 ug/g. This was calculated adjusting for the dilution of standards and also the actual mass of product weighed in. This was reasonable (!) close to the value that had ben measured by the old methoid (which we are trying to superceed) of 4600 ug/g
Now appart from any howlers that others can spot, I have some questions regarding what I've done. The main one is 'How does one choose the concentration of standard to add to the samples'? I repeated this instead adding 2500 ug/g and 5000 ug/g using the same method and proceedure and got an answer of 4800 ug/g ethanol. In fact thats the only question I can think of at the moment. I'm just trying to learn more about how to use this technique as I think it can solve a few other problems that we have (both GC and HPLC methods)