Hi, all -

We recently saw the energy on our Shimadzu RI detector slowly ramp up and max out at 99999. This was coupled with periodic poor chromatography (really crazy-shaped peaks and baseline spikes) that seemed indicative of detector problems.

We changed the lamp, and the energy on the new lamp is slowly creeping up. What is the expected equilibration time for a new lamp in an RI? A day? Does anyone have any thoughts as to what else may be causing this?

Thank you.

H_H
Hi, all -

Our energy eventually leveled off after about 18 hours. The new lamp solved our problems. It is apparent that we had a bad lamp and that a day or so is required to re-equilibrate.

H_H
Thanks for the update! I'll keep that in mind for future reference!
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 19 users online :: 2 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry