Can anyone get me started (and hopefully finished) on the following setup:
We are giving a second life to a HP1050 system which is run by Chemstation. We want to hook up an Alltech 800 ELSD which we also inherited. I am sure that someone out there has done this and it comes down to whether or not they see this post. We have an HP35900 Interface so it comes down to what are the right connections and cabling. I am thinking:
- [Alltech800 analog out] to [analog in on 35900]
- [35900 HPIB out] to [HPIB stack on 1050]
Should I also use the 9 pin remote cables and if so to and from what?
Any hints and suggestions are welcome.
Gabe
Yep, I've got a similar set-up. You don't state whether your 35900 is a 35900D internal circuit board that goes in the computer or a 35900E external box.
35900D: I've never been able to get a 35900D to work with Windows newer than Windows 95, but I'm not so good at computers. Attach a signal cable to your detector. Attach the 9-pin remote from Remote "A" on the 35900D to the 1050 stack (important). Go into ChemStation Configuration Editor and scroll down and add the 35900D module, click Add. Enter your 35900D select code (typical is 6, or whatever you've selected with the dip switches), then click on both "A" and "B", then OK. Save the Configuration.
35900E: I've used with Windows 2000. Piggyback HPIB cable from back of computer to the 35900E. Attach a signal cable to your detector. Attach the 9-pin remote from Remote "A" on the 35900E to the 1050 stack (important). Go into ChemStation Configuration Editor and scroll down and add the 35900E module, click Add. Enter your 35900E select code, then click on both "A" and "B", then OK. Save the Configuration.
If you want to use external event switch to zero detector, that's under "More Injector" and I believe "More Pump" under Instrument, but you'll need the special reddish-brown connectors that go on the back of those modules.