Agilent prep ALS - Waters Fraction collector III problem

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Has anyone been successful in sending a start signal from an Agilent G2260A preparative automatic liquid sampler through a 9-pin remote cable to a Waters Fraction Collector III? I am able to start the fraction collector by shorting out pins 3 and 4 at the back of the unit, however, the start and ground signals from the ALS fail to start it. Thank you for your time.
Hi
You sometimes find the start / stop from the 9 pin is a bit too quick for some other instruments. I believe the G2260A like most 1100 inits will accept an external contacts board, with this you can change the speed of the open and closure which should fix your problem.
Thank you for your quick reply, LCbob - I will try this. All the best, Tony.
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Majestic-12 [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry