Variation in the retention timein different day analysis

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Can any one suggest me how much variation in retention time is acceptable between different day analysis? Suppose today we have analyzed a product and main peak elutes at 10 minute. So when we analyze this product on different day say after one month, how much variation in retention time is acceptable? If there is any guideline then please inform.

Thanks
Sanjiv

In principle, any time window is OK so long as you meet specificity requirements (i.e., no other compounds eluting within that time window).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

As a general rule we are using a criteria of +/- 10%, from the retention time specified in the respective STP, if the retention time window is not specified in the STP.
3 posts Page 1 of 1

Who is online

In total there are 8 users online :: 2 registered, 0 hidden and 6 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider], Semrush [Bot] and 6 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry