GC-MS problem (methanol is converted to formaldehyd?)

Discussions about GC and other "gas phase" separation techniques.

62 posts Page 5 of 5
Newest data

Image
The first thing that strikes me is the inlet pressure - 97.5 KPa pulsed to 241 KPa will give you very fast gas flow rates, and the high pulsed splitless pressure will increase the probability that some solvent condenses at the head of the column and causes peak splitting.

Unless there is a good chromatographic reason for having the pressures this high (and "we have always done it like that" does not qualify as a good chromatographic reason !) I would lower the inlet pressure to give a volume flow through the column of about 1.5 ml/min (on the GC you can set either constant volume flow or constant linear flow, setting the linear flow to a constant 50 cm/s will be the easiest). Then, for troubleshooting purposes at least, leave out the pressure pulse.

All your rention times will be much longer of course.

Peter
Peter Apps
62 posts Page 5 of 5

Who is online

In total there are 64 users online :: 2 registered, 0 hidden and 62 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Georguslat, Google [Bot] and 62 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry