Help: GC-uECD analysis OCPs/PCBs (EPA method 608)

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hello fellow chromatographers,
I have been struggling with a linearity/calibration issue for almost 6 months while analysing organochlorine pesticides (OCP) via EPA method 608 and I seek your advice and assistance. I will try to start from the beginning with as much detail as possible.

I am using a 7-point calibration:
0.009, 0.010, 0.020, 0.040, 0.100, 0.200, and 0.400ug/ml of OCP.
Linearity is assumed by calculating percent relative standard deviation which should be under 10% for this method. This criteria was never difficult to achieve until about six months ago when suddenly the system became non-linear and I started to see this phenomenon:

The response factors steadily Decrease as the concentrations Increase. In the past, the response factors were almost identical over every calibration level.

I am wondering if anybody has ever seen this phenomenon before. All of the tech support people I talk to have never experienced this kind of trend with an ECD.

I have eliminated the following as probable causes of non-linearity:
-The calibrators
-The supply gas purity

I've also tried running a calibration without the guard column, using a different injection port, and using a different detector. So far, none of the troubleshooting has resulted in a calibration as good (linear) as before the I noticed a problem.

Sometimes I suspect that I may have done something to damage the detector during an ECD wipe test. I had performed an ECD wipe test right around the time the non-linearity problems started occuring. But I've looked inside the detector for debris and found none. I also checked the make-up gas fittings for correct distance and column distance into the detector. Also, The baseline readings are low with no negative or ghost peaks.

I have been able to improve the linearity to an acceptable level by increasing the detector temperature and make up gas flow rate. But this feels like a band-aid solution because I don't know why the system changed so drastically to begin with.

Any help or insight would be greatly appreciated. I have an overwhelming amount of detail to provide, but I did not want to overwhelm the main issue for the initial post.

Thank you!
2 things with this:

1. In my lab we would only run as high as 0.2ug/ml and sometimes had to drop that high point because of linearity issues.

2. you only need 3 calibration points for 608.

We hardly ever had samples with actual pesticide hits in them so dropping our high point down to 0.1 or 0.05ug/ml was not a problem. In my experience 608 was difficult with our older instruments and no maintenance ever helped. Maybe just make sure you're using brand new standards.
2 posts Page 1 of 1

Who is online

In total there are 47 users online :: 0 registered, 0 hidden and 47 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 47 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry