Dear Rod, I am aware that the injector temp. was far to high. This was just an example one of the few where I started with injection temp. 70C and for regular measurements I was using 120 but it doesn't change much. Which temp. program would be optimal to resolve air and CO2? I am afraid that with this big tailing it is not possible. Why when I cooled down my sample I could see good resolution of air and CO2? Has water something to do with it? I am not carrying much about the water peak I just want to be sure that I can evaluate amount of CO2 and that water is not damaging my column. I will try what you suggested next week and I will let you know. Do you think my column is fine with this big tailing?
Dear AICMM, I have both loop and syringe, at least I think so

I have an auto-sampler for TCD line which is sampling with the loop. My reactor is connected over vale with GC, during sampling valve switch and fill 500 ul loop and then inject to column. If I sample manually, I am applying additional pressure of carrier gas (I guess this is the make-up gas then) and I am filling the loop by syringe (500ul gas sample). I was also trying with the higher flow rate, the tailing was smaller but the resolution as well.
As you see I tried everything and slowly I am running out of the ideas and that is why I thought that the packed column at least the one I have can not handle with water and therefore I wanted to change to PLOT system. Do you think my Porapak Q is damaged or something? Rod do you think that buying a Chromosorb 102 as you suggested will solve the problem or I need a pre-column anyway? What about argon as a carrier gas?
Sorry for being so annoying but I need to make the system working within two weeks and for the moment I am stucked.
Thanks for help
Kasia