Base line problem using ion pair reagent

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

6 posts Page 1 of 1
I am trying to work with a method that had a great baseline during for months, and now has a bad baseline. I would call the base line wavy, but that would imply that it is regular.

The method uses 1-octanesulfonic acid sodium salt in the mobile phase, and a C18 column. I have used more than one column, different lots of the ion pair, and even different LC's, but the problem is still following me.

Any suggestions on how to get my good baseline back?

Could the temperature at 40 deg C be part of the problem?

Is the baseline change immediate or after the column/mobile phase have been in use for some time? This may simply be a matter of strongly retained junk building up on the column over time, then bleeding off. Try flushing the column with strong solvent (MeOH) and then go back to your normal mobile phase to see if that quiets things down. The temperature shouldn't be a problem. - John

Maygo,

Do you imply that the regulation at 40 C is what changed between the method that worked ((no baseline problem) and now (base line problem)?

If your column temperature control system does not do a good job (i.e. it heats at 40 +/-5 C) then you might have distrurbed equilibrium of your column dynamic modification, leading to increased amounts of ion-pairing eluting out of the column when the temperature increases (i.e. 45 C), then decreasing as the column tries to re-equilibrate at lower temperatures (i.e. 35 C) leading to the "wavy" baseline...

Have hou checked your water source?
Mark Tracy
Senior Chemist
Dionex Corp.
i dont think temperature is creating any problem if it is stable or with in a given range,but most of the time problem with C18 column is washing after use u should try washing it with 50;50(organic:diwate) for a 10 colume volume atlest after use and store c18 column in 80:10(organic:diwater) , and when u start your run make sure u give some equilebrium time for column, hope this will work out
Avi patel
I had cleaned my column and system for hours before and after use to see if that could clear up the problem.

I ended up switching to a different type of LC and (using the same column, mobile phase, and parameters) the baseline was fine.

Either I had been on two LC's that needed further cleaning, or there is enough of a difference between a Water's Alliance, and an HP Agilent to cause the baseline to clear up.

One other possibility is that the Alliances have their detectors on top of the systems with the Air Conditioning blowing on them, while the Agilents have their detectors on the bottom with the lines better shielded from temperature changes.

If I have time, I will run on an Agilent with the tubing going into the detector protected from drafts.
6 posts Page 1 of 1

Who is online

In total there are 22 users online :: 2 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Google Feedfetcher and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry