Regeneration and cleaning steps depend on the polarity of the impurity. Flow rate show be 25 - 50% of normal working flow.
POLAR IMPURITIES
Rinse out buffer solutions with 80 - 20 water: organic solvent. 20 column volumes.
Rinse with a mixture of high concentration of organic solvent 70:30 to 90:10 methanol: water or acetonitrile:water for at least 10 column volumes.
Most polar impurities are eluted when the buffer is rinsed out.
NONPOLAR IMPURITIES
Rinse column with a high water water ratio in order to remove buffer. NOTE do not go over 90% water in order to prevent phase collapse.
Rinse column with 100% methanol to remove polar compounds organic compounds
Rinse the column with 100% acetonitrile to remove medium polar organic compounds (temperature can be increrased to 40deg C)
Rinse column with THF to remove nonpolar organic compounds
All rinsing should be at least 10 column volumes
This and other helpful hints are found in Macherey-Nagel useful APPLICATION GUIDE book. For a FREE book go to