Dear All,

I am observing a steep negative baseline drift as the concentration of acetonitrile rise in my gradient.
the method details are described below:

Mobile Phase A: Acetonitrile/25mM Phosphate Buffer pH 7.0 27.5:72.5 v/v
Mobile Phase B: Acetonitrile/25mM Phosphate Buffer pH 7.0 60:40 v/v
Standard Diluent: 27.5% Acetonitrile 72.5% 25mM Phosphate Buffer pH 3.0
Column: Luna C18 (2), 5µm, 250 mm x 4.6 mm.
Guard Column: C18 4 x 3.0mm cartridge contained in Security Guard  Universal Holder
Injection Volume: 100µL
Detection: UV at 239 nm
Flow Rate: 0.75 ml/minute
Column Temperature: 40°C
Gradient: Time, min. % A % B
0 100 0
30 60 40
30.1 0 100
44 0 100
44.1 100 0
52 100 0
I conditioned the column by running 3 balnks before starting my analysis, but the baseline negative drop is still there. I do not know the column history as this was supplied by a customer, and in their chromatogram the baseline look flat. I would expect the baseline to show a positive drift when the % of acetonitrile increase.
I would much appreciate if you could give me advice on why I observe this drift and suggestion on how to solve this issue.

Many Thanks

Emanuele Scollo
Try removing the column and replacing it with a meter or so of capillary tubing (0.1mm id), then run the same gradient.

If you see about the same amount of drift, then it's obviously not a column problem, and I would suspect that the flow cell in your detector is out of alignment.

If the baseline looks OK without the column, then it obviously *is* a column problem. Washing the column with 100% acetonitrile may clean it up, but if it is heavily contaminated, it may be more cost effective to simple replace it.

We put together a brief tutorial on gradient baseline issues (taken from our HPLC Basics, Equipment, and Troubleshooting course). You can view it on the LC Resources web site:
http://tinyurl.com/47gwd4c
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 39 users online :: 1 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Ahrefs [Bot] and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry