I have a very basic question: While developing an RP-HPLC method by increasing an aqueous phase percent in mobile phase, what solvent should be used to prepare sample? Richer in organic or aqueous phase?
Anna Andrzejewska-Santiso
In general, the diluent should be the same as the mobile phase. When that is not possible or practical, more aqueous is better.