Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
For an analytical method we have set a criteria of an RSD of 2.0% or less for the repeatability of six standard areas at the start of the run and a limit of 2.5% or less for all of the run.

We typically see values of 0.5% and 1.0% for the first six and all standards respectively and over 50+ runs have not exceed 1.0 and 2.0%, should I reduce the criteria in respect of what we areactually obtaining or will I be creating a rod for my own back and increasing the risk of failure?

Thanks
Chris

What we do -

For dilute samples (dissolution samples of things with weak chromophores, low area counts) - we go with a 5% limit on standard RSD for initial standards and we stick with it for the run. If RSD strays above 5%, we'll try nested bracketing and reinject any sample sets not bounded by standards that meet the 5% criterion.

For more concentrated samples, it is as above but with a 2% criterion.
Thanks,
DR
Image

Also, do you think a target retention time of 9.0 to 12.0 minutes is too wide?

We do not set windows in our methods. We specify that the chromatograms look approximately like the example(s) provided. We also typically specify a minimum resolution between whatever the critical peak pair is for a given method.
Thanks,
DR
Image
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 2 registered, 0 hidden and 14 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot], Bing [Bot] and 14 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry