I used to have similar problems with getting the BFB to pass. We've moved our GC-MS to a new lab recently and I've had no issues since. The new lab is much cooler than the old one and this might have helped. Have you looked at storing the BFB solution in the fridge between uses?
-
- Posts: 22
- Joined: Fri Dec 11, 2009 3:55 pm
What kinda of source do you have? Inert? Non-inert? and who makes it? Agilent? EST? We've recently bought an EST inert source and their specs for use didn't work with our instruments to well so much so that mass 175 wasn't showing up either.
So we set everything back to what our Agilent source was running at, ran a manual tune on it and 175 showed back up.
So we set everything back to what our Agilent source was running at, ran a manual tune on it and 175 showed back up.
17 posts
Page 2 of 2
Who is online
In total there are 103 users online :: 1 registered, 0 hidden and 102 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 102 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Ahrefs [Bot] and 102 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science