RSD of 5 injections is 0.1%.
First Std Injection Plate count: 17198; Tailing: 0.94
Bracketing Std Plate count (After 15 injections): 14593; Tailing: 0.94
Plate count decreased due to the increased peak width: yes all the peaks show same problem.
All my Initial runs have shown plate count about 72000 & Tailing about 1.4.
Sounds like the inlet frit is fouled or the column is voided.
Have you tried a strong organic flush on this column?
If not, purge the column with 20 column volumes (1CV ~ 1.5mL) of 10/90 ACN/H2O to remove all buffer. Then run a gradient from 10/90 to 100% ACN over 10min @ 1mL/min and hold at 100% ACN for 10 CV. Next flush the column with 5 CV of 100% IPA at a low flow rate - be careful not to over pressurize the column as IPA gives VERY high back pressure. Next flush with 20 CV 20/80 MeOH/CH2Cl2. Flush again with 10 CV 100% IPA to remove all CH2Cl2 then flush with 5 CV 10/90 ACN/H2O.
Equilibrate with 10 CV inital mobile phase and see if perfomance has improved.