My latest findings regarding the injector of the Acquity autosampler is disturbing enough to earn an own post. This could affect many people.
Situation 1. Partial loop, needle overfill injection (50 µl loop, 250 µl syringe). Injection volume 20 µl.
Injection from a 1 ml vial "as is": main peak area 3670000
Addition of 200 µl ACN to the same vial: main peak area: 3520000
This caused a lot of turbulence since it clearly indicated that we had a adsoption problem of the product. We should see about 83% of the area in injection 2, but we saw 96%. Acetonitrile increases the response.
Situation 2. Full loop injection (50 µl loop, 250 µl syringe)
Injection from a 1 ml vial "as is": main peak area 9010000
Addition of 200 µl ACN to the same vial: main peak area: 7780000
Now this is more how it should look! I cannot see any other explaination than that the Acquity in partial loop mode injects a larger volume when the sample contains acetonitrile, compared to the normal sample. In full loop, the sample solvent makes no difference. (I have repeated these experiements many times, with the same result)
Since it is often is not possible (or desired) to prefectly match the solvent composition of the standard and samples, my conclusion is that full loop must always be used if the Acquity is to be used for quantitative analysis.