krickos » Fri Oct 08, 2010 6:21 am
Hi
I see a number of potential pitfalls with this analysis, apart from triethylamine being sensitive to active sites in liner or on column (however less likely on a 624 column with a thickness of 1,8-3µm).
First the injection port temperature seems derived from some default headpsace injection, it is too low for liquid injections as DMSO has a boiling point of ~189°C.
Second. Actetates (like ethylacetate) can undergo both basic and acidic hydrolysis, then we have potential matrix issue between API and the amine/acetate aswell. Consequently pH control may be needed and and standard addition calibration is strongly recommended.
Third. Column oven isothermal at 200°C. This is obviously not good from a resolution point of view/peak shape aswell. I would rather start at 40°C for a few minutes and then ramp up with 10°C/min to start with.According to litterature data this approach should resolve all you solvents nicely on that particular column type.
Fourth. Split ratio. Again it seems derived from a headspace method for residual solvent testing. Even if youi now increase injection tport temperature, the realtive huge amount of DMSO on column may smear out you peaks.
Other: I would do a standard first with only DMSO/triethylamine and see if I could get that working first. As mentioned DMSO might not be suiteble another solvent might be needed as DMF, DMI or in worst case N-methylpyrrolidone as mentioned by Kalidass. Kalidass also seems to have a headspace injection procedure that works so changing to headspace injection may also need considering.