Changing a detector temperature sensor (ECD)
Discussions about GC and other "gas phase" separation techniques.
19 posts
Page 2 of 2
John, did you remedy this problem? I am having the same problems with my 5890 Series II gc.
-
- Posts: 170
- Joined: Thu Mar 30, 2006 9:35 pm
Yes, I did.John, did you remedy this problem? I am having the same problems with my 5890 Series II gc.
From my maintenance log:
Started analysis method and got this error:
FAULT: DET A TEMP RDG
All heating zones seemed to be defective; oven, injectors, and detectors. Oven reading 824, detector reading -200.
Replaced main board upon suggestions from chromforum.com. Was still getting similar errors after replacing main board.
Replaced oven temperature sensor only and instrument started functioning as needed...? Not sure how one bad sensor could make all others stop working. Regardless, this worked and was not very difficult.
Hope this helps!
John
-
- Posts: 5
- Joined: Fri Dec 18, 2009 5:53 pm
I had no luck with replacing only the board. =/
-
- Posts: 17
- Joined: Wed Aug 25, 2010 2:35 pm
Hello,
I am working on the same GC and got the same message (FAULT:TEMP RDG DETA) and after replacing the oven sensor my temperature is not reading correct. I am getting the Actual -200 deg and set 50 deg which is incorrect reading. I am not even using Detector A and not sure why the original message appeared. However the fault message is gone but the problem still exist. Please help. Thanks
I am working on the same GC and got the same message (FAULT:TEMP RDG DETA) and after replacing the oven sensor my temperature is not reading correct. I am getting the Actual -200 deg and set 50 deg which is incorrect reading. I am not even using Detector A and not sure why the original message appeared. However the fault message is gone but the problem still exist. Please help. Thanks
19 posts
Page 2 of 2
Who is online
In total there are 110 users online :: 0 registered, 0 hidden and 110 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 110 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: No registered users and 110 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science