by
osten » Wed May 18, 2005 5:39 pm
Many thanks to everybody who has answered.
To MG: Yes, I use Agilent 1100 MSD. Optimal Vcap and Frag were determined in FIA. But I've tried others as well. Frag I can change, no problem. At V cap 3500 or 4000 response much worse, but base line is smooth (for 4-5 runs). I don’t see peak m/z 93, but I see in baseline 173, 475 and much of them.
To SIELC_Tech: I’ve seen this advertising. Seems too much NH4OAc (75 mM). TFA I’ve tried instead AcOH. No signal at all, including while direct injection in chamber.
To Steve Reimer: I really do know. I think the both ones. I‘ve made injection without column and mass response was several times lower then for the same concentration of two other compounds of similar (?) structure: mepiquat (N,N-dimethyl piperidin) and chlormequat (2-Cl-Ethyl-trimethyl-ammonium). The samples were dissolved in H2O or in H2O/ AcOH/ NH4OAc.
Concerning the chromatography. The quter. amines chlomequat and mepiquat in the same conditions have passed the column (perfluoroPh) without problems. In the case of PQ and DQ I see either nothing or in the best case unclear signal with tailing 1km. I’ve read about the sorption in glass, so I used different concentrations - no visible improvement. May be the problem in structure: bipyridilium salts and aliphatic and their interaction with column?
Thank you