Hello, I can sympathise with your problem as have experienced it myself (several times, seem to be constantly having to work near the assay LOQ, sigh!)
As Uwe says, things you can do to increase assay sensitivity are to;
1) Concentrate your sample during extraction
2) Use a micro-bore column (e.g. 1mm i.d.)
3) Inject more sample on to the column
The concentration factor that you can achieve will be dependent on how much blood you can take from your patients/volunteers (which will in turn be dependent on how many samples you want to take and the type of patient (baby, child, adult)). Also, it will be influenced by the recovery of your compound as you may find that recovery decreases as you attempt to concentrate your sample more, so you will have to experiment to find the optimal conditions. You also have to consider your required injection volume at this point; you will need to allow for at least two injections for analysis (preferably more in case of problems) and wastage due to the injection process. Obviously, the more you concentrate, the smaller your total sample volume will be.
Micro-bore columns are an excellent way to achieve a fairly large increase in sensitivity. However, you will need to scale your flow rate accordingly and may need to buy a special micro-flow pump if you don't already have one. Typical flow rate for a 1mm bore column is only 0.05mL/min.
As for your detector, there are a few things to consider here too.
Is it performing optimally? (is the lamp old? Has the wavelength calibration been checked lately?)
Flow-cell - are you able to replace the flow cell to enhance sensitivity (go for small volume but longer pathlength)
Are you working at the lamba max (I know the paper you've referenced says 267nm but have you checked this on your system)?
First work out what your current LOD/LOQ is from simple injections of standard solutions, then you'll be in a better position to judge what measures you need to take to achieve your target LOQ.
Best of luck, Ann