Hello,

I am seeking help from you all regarding mass spectra I got.

I have MS spectra of my drug donepezil HCl from Electron Impact and also from Time of Flight analyzer with Electrospray.

But spectral pattern is different.

In EI spectrum i am getting fragmentation pattern while in Electrospary with TOF analyzer there is no such fragmentation pattern only a molecular ion peak was seen at m/z=380 (molecular wt of donepezil) and nothing else.

Plz help me in this regards,

That is normal. Electrospray gives less fragmentation than EI, and sometimes no fragmentation at all. The idea of running the sample by ESI-TOF is to get accurate mass on the molecular ion, to confirm the molecular formula. On the other hand, with EI, you match the fragmentation pattern against a library.

The mass spec operator should be able to generate fragments in the ESI spectrum, if you really want to see them. But, the fragmentation pattern probably won't be the same as in EI.
2 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry