We manufacture cosmetic lotions containing methylparaben and propylparaben. We are getting ready to validate an HPLC method for paraben analysis. I would like to know if there is any recommendation for the specification on accuracy. Currently during preliminary testing, I am getting recoveries in the range of 94% to 103%. I have tried different columns, different extraction methods and cannot improve the recoveries. Please let me know if there is a standard industry guidance on the Specification such as +/- 4.0% or 10.0% etc for preservative accuracy. Thanks a million for you reply.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
US: we also do parabens on consumer lotions, but our precision/accuracy/recovery is much better. Are you incorporating an acid such acetic acid in your mobile phase? We dissolve the samples in methanol and filter, then HPLC.
Here's what we obtained:
"Mean recoveries from regular scent placebo product spiked with three different known levels of each paraben (bracketing current use levels) were 99.0% with a standard deviation of 2.0 for methyl paraben and 100.0 with a standard deviation of 2.1 for propyl paraben with column choice #2 and 100.0% with a standard deviation of 2.1 for methyl paraben and 100.6 with a standard deviation of 1.9 for propyl paraben with column choice # 3".
Here's what we obtained:
"Mean recoveries from regular scent placebo product spiked with three different known levels of each paraben (bracketing current use levels) were 99.0% with a standard deviation of 2.0 for methyl paraben and 100.0 with a standard deviation of 2.1 for propyl paraben with column choice #2 and 100.0% with a standard deviation of 2.1 for methyl paraben and 100.6 with a standard deviation of 1.9 for propyl paraben with column choice # 3".
-
- Posts: 10
- Joined: Tue Mar 15, 2005 7:16 pm
Can you give more details on Column#2 and #3. we use Acetonitrile for extractions. The mobile phase is 50:50 0.1% H3PO4 and Acetonitrile and column we use is C18, micro-Bondapak from Waters.
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Column 2 is a 2.1mm id type A RP-18 silica, 10 cm long with a shorter 3 cm cartridge of same as guard column. Column 3 is a modern "intrinsically base-deactivated" RP-18 sold by major vendors such as Thermo Keystone. The question is: are your peak shapes OK, and if so, is placebo product flat in the regions of interest? These parabens can also be determined by GC, but I'd trimethylsilyl derivatize first for that.
4 posts
Page 1 of 1
Who is online
In total there are 12 users online :: 2 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Google [Bot] and 10 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Amazon [Bot], Google [Bot] and 10 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science