thanks
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
are you saying the peak width of 10% and 50% std are the same if the method is validated and not overload?If are not overloading, the peak width at half height should be approximately independent of the amount injected. And, if the width *does* increase with increasing amount, then you *are* overloading.
Yes, the peak width should be independent of concentration so long as you are not overloading. The peak height will be proportional to the amount injected. That is why peak height can be used for quantitation; width is constant, so area is proportional to height.
The peak half width was different from 1 to 120% std. Their response factors are same using peak area/conc.Yes, the peak width should be independent of concentration so long as you are not overloading. The peak height will be proportional to the amount injected. That is why peak height can be used for quantitation; width is constant, so area is proportional to height.
I will look into my data again, will confirm it tommorrow.
Thanks, it helps.