plate number calculation

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

11 posts Page 1 of 1
If I use diluted sample which gave lower peak height but narrow half peak width, the plate number will be higher. Are I manipulating data?

thanks :oops:

Are you using a validated method and are you referring to diluting a standard as opposed to a sample? I am used to performing plate counts for system suitability checks using standards only (with a known purity and concentration) as samples can be very variable.

same std, but inject less which leads to smaller, narrower peak.
method used for calculating Plates should be fixed. no change including std conc, injectin volume.

If you are not overloading, the peak width at half height should be approximately independent of the amount injected. And, if the width *does* increase with increasing amount, then you *are* overloading.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

If are not overloading, the peak width at half height should be approximately independent of the amount injected. And, if the width *does* increase with increasing amount, then you *are* overloading.
are you saying the peak width of 10% and 50% std are the same if the method is validated and not overload?

Yes, the peak width should be independent of concentration so long as you are not overloading. The peak height will be proportional to the amount injected. That is why peak height can be used for quantitation; width is constant, so area is proportional to height.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Yes, the peak width should be independent of concentration so long as you are not overloading. The peak height will be proportional to the amount injected. That is why peak height can be used for quantitation; width is constant, so area is proportional to height.

I will look into my data again, will confirm it tommorrow.

Thanks, it helps.

Yes, the peak width should be independent of concentration so long as you are not overloading. The peak height will be proportional to the amount injected. That is why peak height can be used for quantitation; width is constant, so area is proportional to height.

I will look into my data again, will confirm it tommorrow.

Thanks, it helps.
The peak half width was different from 1 to 120% std. Their response factors are same using peak area/conc.

That means that your system is overloaded (as far as the chromatography goes). While that is not the ideal situation, it is not necessarily a problem. I does mean that plate number and resolution measurements should be made under "worst-case" conditions (i.e., with the largest standard in the linear range).
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

One can also suspect that the halfwidth measurements of the lower peaks are off. If no mistakes, like overloading, are perpetrated than all peaks should overlap almost perfectly if you normalize their hights to be the same in an overlay mode of the PC software.

This can also happen if the large peaks are out of the linear range of the detector - the top of the peak is blunted, the peak apex is lower, and the half height is lower down the sides of the peak where the width is greater. This shows up clearly if you follow Hans' normalize and overlay suggestion.

Peter
Peter Apps
11 posts Page 1 of 1

Who is online

In total there are 59 users online :: 2 registered, 0 hidden and 57 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Google [Bot], Semrush [Bot] and 57 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry