Totalchrom error question
I have Totalchrom Navigator v. 6.2.0.0.0:B27. I'm trying to get it to communicate with an NCI 902 interface. I was able to get the software to see the interface and set it up in the configuration menu.
The problem is, the system won't initialize. It tries to, but then I get a message that says "port comm error." I can't find in any of the manuals the possible causes for this. Does anyone know?
Thanks!
Chris
Hi Chris,
Have you been using it in this configuration for a while - or have you just started to set this up?
I get this message reasonably often usually when an instrument has a fault. So I turn the instrument off and on. I also close TC and then re-open it. This mostly works for me.
Hope this helps.
WK
If you go into the system configuration verify that the NCI902 has EPROM version 3.4. The early versions of these A/D boxes that Perkin Elmer released needed an EPROM update because of network / communication drop issues. I do not know if this is the cause of your issue but I found this out the hard way. I could get my systems to start up but I would randomly loose my run at some point with a port com error.