Hi,

I hope you can help me with my assay query. I am trying to set up an assay using LCMS-ESI detection. The active component is Ephedrine and I am using Phenylpropanolamine as my internal std. I am just puzzled why I obtained two peaks for each component when I run it in LCMS. I only expected one peak for each component. Could you please explain why? I am a new LCMS user.

Thanks

maybe your injectionvolume is to high? try injecting smaller volume.

...or your injection solvent is too strong for your mobile phase. What is your injection solvent and injection volume when this happens? Also what are your LC conditions (column dimensions, mobile phase composition, flow rate, etc.)
My injection solvent is 0.1% Formic acid. I am using a Hypersil phenyl column, 2.1x100mm, mobile phase is 18%Acetonitrile in water containing 20mM ammonium acetate. The flow rate is 0.3ml/min. Ephedrine produced two peaks and so did my internal standard. I also tried using 18% ACN as solvent for standard and internal std, but I still got the same results. Please advice.

Thanks

Why are you dissolving the sample in formic acid whenthe chromatography is done at another pH? I would dissolve the sample in the same sovent used as mobile phase.

Eliza, I didn't see your injection volume listed. Another way to see if this is an injection solvent mismatch (due to the pH difference) is to try injecting a small volume. For that column, 1 or 2 microliters should minimize or eliminate injection solvent issues.

eliza, what pH is your ammonium acetate? A bad buffer + high concentration of an acid injection can give severe mismatch even at very low injection volumes.
Hi,

My injection volume is 10ul and my mobile phase pH is about 4.8.

Thanks

The other possibility is that the column is dead. If both of your double peaks look similar, than this could be the problem.

I tried to follow the suggestion of choosing the sample extract solvent to be same pH and same buffer concentration as the mobile phase. There was an improvement. However, I also thought that perhaps my drugs are not very stable in solution itself. So, I added a small amount of reducing substance into the sample/standard mixture prior to SPE extraction. It worked!! I only have 1 peak for each component now.

Thank you all for your helpful suggestions
10 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry