GCs that are old enough to vote might be a bit of a challenge to keep running. Kudos to you for getting a full set up before you started, and to your local P-E suppliers for being willing to support them. You did get new columns right ?
Peter
Yes we bought new columns as well. SP-2560 for the cis/trans and a Restek RTX-1 (7.5m X .25 X .25um) for fast chromatography (in-process samples).
My supervisor is considering new GCs (Agilent).

But still, we need to try to run those old GCs.
We ran through a lot of problems, like the "port comm error" caused by the second CPU of the computer. I've disabled it, and it solved the problem.
The guy from P-E didn't plan to work with that old equipement, he was a little bit discouraged.
You don´t have a reference chromatogram for this standard? Do you know what is supposed to be in it? Looks to me as if you have enormous amounts of dirt/decomposition. How many times did you do this? Did you do what Peter suggested, namely, inject less?
The standard we are using come from a similir plant in Pittsburgh (we are based in Canada). The supervisor from this plant is here to help us to mount the lab. I'll try to get the real chromatograph.
Yes the sample (std) is old (maybe 2-3 weeks) and there is a lot of decomposition, I agree. We are just trying to get good resolution and no more peaks tailing before going with another standard, and then, some samples.
This morning, I tried what
mahdiebrahimi told me. We are getting better elution but still, need some improvements. I increase the Bunching FActor to 4 and noise threshold to 10.
Next step, inject less (0,5uL instead of 1,0 uL). If there is no change, i'm gonna dilute the sample. I'll post a new chromato when the inject less step will be done. Finally, if this isn't enough, we will have to wait for the Flowmeter to be delivered... I borrowed one from another plant near us to set up the GC with the SP-2560 column.
BTW, thanks all to support me working those GCs... Their giving me some headaches
EDIT: Might reconsider changing the seryngue. The one we are using isn't Gas Tight. I increased the injector/detector temp to 350 C to flush.
When i'm injecting hexane only, there is a lot of unknown peaks (suppose to be pure hexane). I'll check this.