i prepared my new fresh mobile phase and i still see the negative peak at that concentration ( 0.0005 down to 0.0001 ppm ) but if i go up than 0.0005 ( say 0.001 i don't see negative peak at this concentration ) !
Then your new batch of mobile phase is also contaminated.
What do you see if you inject a deionized water blank (I'm assuming reversed-phase chromatography here)?
My reason for suspecting contamination is that the absence of something chromatographs in the same manner as the presence of that same compound (the phenomenon has been exploited under the name "vacancy chromatography").
Many years ago when I was working with single-column ion chromatography, we used Santa Clara, California drinking water as a quick-and-dirty check sample (it contained ppm-level amounts of chloride, nitrate, and sulfate). Just for fun one day we prepared our mobile phase buffer with tap water and injected DI water. We obtained a chromatogram with *negative* peaks at the correct retention times for those ions.