Hi everyone,
Has anyone tried to adapt the residual solvents method in USP- 467 to use a liquid autosampler? If it could be done, would it require full validation? What injection parameters would you recommend?
Thanks!

USP <467> allows liquid injection or headspace for sample introduction, so many people have used an autosampler for this method. Start with autosampler defaults and a 1 uL injection, and change parameters if the defaults don't work.

Hi

If doing liquid injection you may need to keep an eye on injection temperature even before you start depending on what sample solvent (water,DMF etc) and boiling points of residual solvents you do analyse. If memory serves typical default injection temperture is like 140°C that may be too low.
Split ratio may also be tweak to optimise sensitivity.

If using water as sample solvent you likely need to reduce injection volume immediately to like 0,5µl for a ~1ml liner.
3 posts Page 1 of 1

Who is online

In total there are 52 users online :: 1 registered, 0 hidden and 51 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot] and 51 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry