Sorry, but Ihave to disagree with a lot of what dhe0124 says.
Downward drifting baselines are certainly not common unless there is something wrong with the system. Upward drift (due most commonly to bleed) is inevitable to some extent in programmed GC.
I cannot imagine any scenario in which doing all this preventive maintenance before every run could be worthwhile.
The 1/8 inch insertion depth for the column is specific to Agilent instruments. Cutting 1/8 inch off the column is pointless - contaminants will penetrate at least 50 cm. You should though cut a cm or so off the end of the column after it has been threaded through the ferrule.
I have never heard of pre-conditioning vespel-graphite ferrules, and I have never experienced their shrinking with time once they have bedded in over the first couple of runs and been re-tightened. I have had VG ferrules in place for months and hundreds of runs with no long-term shrinkage.
Leak checking is good practise, but never ever with Snoop. Get a good leak seeker, or if you want to use liquids use propanol diluted with is own volume of water. NB when using a leak seeker turn the oven fan off. If you use liquids you have to cool the inlet below their boiling point.
Baking columns near their maximum temperature shortens their lifetime. It may be helpful, but it should be invoked only if the samples contain a lot of heavy contaminants. Baking is a lot more effective if the flow through the column is reversed. Paradoxically baking out, especially for long periods can make upward baseline drift even worse by damaging the stationary phase, which then starts to bleed more.
Measuring flow rates is good practise, but is only really valuable if it is done with an external flow meter - the front panel readouts of the GC are linked to its gas controls, and so a fault in a sensor can cause an incorrect flow with the expected indicated flow still displayed on the front panel.
FIDs (especially with capillary columns) last months or years without collecting any debris. The only significant sources of contamination are large volume injections of chlorinated solvents (which make soot), and silylating agents which make silica. It is certinaly not neccessary to clean an FID (which usually implies cooling it, dismantling it, reassembling, heating up and conditioning) before every run, - you would maybe get one analytical run done before it was time for column's overnight bakeout !
Welcome to the forum by the way.
Peter