Hi there, I did inject a 100 ppm of methyl estrate with scan mode I got an area of 409003309 and with SIM mode I got 132228767. Can some one tell me if my GC MS has a good sensitivity. Faris
-
- Posts: 1680
- Joined: Sat Aug 23, 2008 12:04 am
The sensitivity specifiction for GC/MS will be a signal to noise measurement of a few picograms on column, with a specific column, flow rate, temperature profile and acquisition rate. Also make a blank solvent inejction between injections of high concentration solutions of ethyl sterate and low concentration solution. Sterates tend to carry over between injections.
Eyeball estimate, you have a shot at a visible peak at 10 ppb and even 1 ppb. So as long as the noise is low, you may be in the ball park.
Eyeball estimate, you have a shot at a visible peak at 10 ppb and even 1 ppb. So as long as the noise is low, you may be in the ball park.
2 posts
Page 1 of 1
Who is online
In total there are 50 users online :: 1 registered, 0 hidden and 49 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 49 guests
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm
Users browsing this forum: Bing [Bot] and 49 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science