Is there any way irreversible aggregation can occur within the column without a permanent increase in back-pressure?
That was my idea too and I think that’s what you’re dealing with. The peptide (or some of it) aggregates due to the low pH and high ACN content in the mobile phase. The aggregates stick on the stationary phase and slowly dissociate and elute over time.
To test this thesis, try and inject a blank (DMSO containing solvent) right after a normal sample elution/run. Maybe you’ll see a peak representing some of the peptide that you didn’t recover in the previous run.
If you’re comfortable with the conclusion you can go on and try to optimize the conditions in order to avoid the sample loss you’re experiencing at present.
Here are two suggestions I’d explore if I were in your situation:
1. Choose less hydrophobic stationary phase (column) e.g. C4 or even Phenyl.
2. Change mobile phase e.g. some diluted buffer approx. 0.01 M and alkali pH e.g. 8 – 9 if the column material allows for it.
Remember to test the solubility of your peptide in the modified/new mobile phase, if you choose to follow that path. The way to do it is, mixing some of your sample solution with the mobile phase in a test tube and observing the mixture for cloudiness. Actually it’s also a relevant test to carry out with your current mobile phase.
Best Regards