Force Baseline by Time(with no value) and with value events
I am confused with these two events described in waters manual. The description is not clear to me.
It says baseline is drawn from event start to event end. But in the figure it shows the baseline is draw from a point after event start and to a point before event end.
Also I could not digest the concept of baseline averaging. What do you mean by baseline averaging and also how do you set baseline points based on baseline averaging.
Can someone help.
Hi sana_sankar,
I have a little spare time (thanks to an inconsiderate volcano in Iceland! and hence all my posts today...) and I just noticed your post has not been answered.
I hope I can clarify the Force Baseline by Time event in Empowers Traditional integration.
With Force Baselines by Time event, the baseline is drawn from the event start time to the event end time.
I believe your initial confusion comes from looking at the Force Baselines by Peak event.
For this related Force Baselines by Peak event (but not the same event), Empower examines the time region between the event start and end times. The base line is then drawn from the peak start of the first peak after the event start time, to the peak end of the last peak prior to the event end time. Therefore the Force Baselines by Peak event appears to “show(s) the baseline is draw from a point after event start and to a point before event endâ€
Empower
MVC - Medicon Valley Consult
www.mvc-consult.com
In total there are 16 users online :: 0 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Users browsing this forum: No registered users and 16 guests