Labgurl » Fri Apr 09, 2010 11:38 pm
I am testing for Hydrocarbon content in condensate petroleum using 7890A, TCD, Packed column. We recently replaced our column, and had to make some adjustments to the injector & TCD temperatures. The run is fairly short....methane, ethane, propane, iso & n-butane and iso & n-pentane all elute in about 4 mins and then there is a valve switch on and the backflush is started, in which hexane+ is all summed as one peak, and the backflush valve switch off is at about 12mins. It is every time after the backflush is complete and the valve switches off that another peak appears, which we hadn't seen before. I have tried increasing the run times, changing the valve switch times, adding another valve switch, and nothing seems to work.....each time after the last valve switch off the mystery peak appears.
When samples are run in sequence, this mystery peak from the first sample carries into the inital (methane, ethane, propane) peaks of the next sample run.
Could some sample be condensing or be trapped in the valve area?
Injector Temp: 200oC
TCD Temp: 250oC
Column Temp: 100 oC
Valve Temp: 125oC
I have tried baking things out.
Any advice?
Thanks
Labgurl