- Holly
I'm experiencing problems on extracting vitamin D from calcium tablets. Does anyone have a solution? I'm also having problems on filtering the samples. I rather follow the BP method, by reversed phase chromatography.
Thanks in advance.
Holly
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Have you made sur that you use sun protected glass?
vitamin D is affected by sun light and heat.
make sure that your samples in the HPLC are in a cooles autosampler tray. at least 15 degrees. better go to 10 if you can, although not too much under that because i have seen it also deposit incide the vials after a while if cooled to close to 4-5 degrees.
do you get any problems when injecting a standard solution? like RSD problems?