Triple Quad Quality Wich brand

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
We are just using a Thermo Quantum Ultra since 6 month. It seems, that the machine is unstable in the masses >1000 d.
Not easy to find out, but now also the technician beleves it.
Does anyone of you have similar experiances with Thermo?
How do you consider the general quality of these machines?
Thanks
math

Math,

We have 8 Thermo ion traps, one LTQ-FT-MS (ion-trap-FT) and one old TSQ. They all perform pretty well.

Your problem sounds a little bit odd, but it can be an isolated problem with the instrument you have. Is the technician, someone from Thermo or is he someone that works in your labs and just operates your MS?

The technician is from thermo.
Before they thought, its a substance specific problem. But now, they saw, that the calibration (of the machine MW etc.) is worse even after one week. So it seems a fault of the machine. Lower MW stuff works well so far.
3 posts Page 1 of 1

Who is online

In total there are 43 users online :: 1 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry