lmh » Fri Jan 15, 2010 12:01 pm
From your comment of Agilent1100 I assume you are using LC-chemstation.
The sequence parameters window allows you to choose a path from a list of paths you have defined (at least in Revision B). To define them, Revision B onwards, stay in the Method and Run view, and select "View" - "Preferences" from the menus at the top of the screen. This brings up the data and method file paths, with "add" and "remove" buttons so you can add a new path. This will then become available on the Sequence Parameter page.
In older versions of Chemstation, the easiest way to change the path is to close chemstation, open the Configuration Editor, select "Paths", choose your new path, save the configuration, and exit, and re-open chemstation. There is a way to do it from the command line at the bottom of the screen, but I can't remember the thing to type.
On our Agilent 1100 system, if you run a sequence with "STANDBY" selected in the post-sequence macro place, then it turns off the pumps and lights in the DAD, but it leaves the autosampler thermostat on cold, so if yours is the same, you don't need a macro.