We use auto sampler with pull-to-fill injector. These days we have trouble with poor precision reproducibility in series injections from one vial.
We always fill the vial up to ¾ capacities.
The septa are bought from two suppliers. The original septa for the LC are thinner than the other and more fragile. In 2-3 injections it will be broken into 2 parts and down to the sample solution inside the vial.
Then we use other septa that be used in different LC. These septa are thick and more elastic; do not be broken after repeated injections. But new problem appears: poor precision.
After we inspected any possible reasons, we come in a conclusion that the cause of the problem is vacuum cavity in the vial after injection, so that the plunger can not draw the solution perfectly.
We tried to open the vial just a moment before injections, and we got good precision.
We tried to use the original septa for the LC, and we got good precision … and broken septa in the sample solution.
We tried to use the septa from other LC, and we got bad precision!
The question:
Is it safe to use the original septa that always be broken after several injections?
We just worry it will contaminate the solution and the debris can block the needle and the line.
What should we do to eliminate vacuum cavity when we use the other septa? Should we open-close the sample vial after each injection (it works and gives good precision)?