I’ve been giving it some thought....
And just reading the above responses:
It does sound as if you have a hardware communication issue.
The GC should only start an injection when it is triggered by the headspace.
But if the HS injects while the GC is acquiring, the sample goes straight out the vent, as you see by no mid-run spikes.
And your HS should be smart enough to be able to take the vials into the oven at the correct intervals to be pressurised and ready to inject after the GC has had time to come back to initial temp (cycle time).
And it should also wait for the GC 'Ready' signal before it injects. As Peter pointed out the headspace will wait with the sample in the injector loop, pressurised, until the GC is ready.
So, it sounds to me as if your headspace is not waiting for the GC to come ready before it injects. This is why extending the GC cycle time setting for the headspace has solved (masked) your problem. As Krickos said, the cycle time should be around 38 minutes (with 5 min safety margin).
- So you're right that the method is incorrect, but not that the thermostatting time is incorrect, but that the cycle time is too short.
(And to be realistic, going from 220C to 40C in 5 minutes is do-able, so a cycle time of 25 min will probably be OK in that the sample wouldn't be sitting in the loop all that long before being injected)
So, your injection should have waited for the GC 'Ready' with the sample in the loop.
But, it didn't.
Hang on.... Just noticed it also didn’t thermostat your subsequent vials for the correct time, so maybe you also have a firmware issue with your HS as well.
All signs point to your HS.
Reboot everthing.
Double check you set up. Try again.
Oh, and a bit more info could be helpful.
How long have you been running this method, is it something new, or new hardware or software?
Does the system work for other analysis?
Oh, and not dissin’ . Just trying to help with the troubleshooting!
Let us know how it goes.