I just connected an old fluorescence detector but what it gives me were spikes. Was my connection wrong or the cell/lamp have problems?

Assuming that the lamp started:
Did you have a flow through the cell while observing spikes? If so, stop the flow and see what happens, or flush the cell with de-airated water and check under stop flow, make sure there is no air in the cell. Also, keep the detector on for two hours before doing anything with it. If the spiking is small enough to get a spectrum I would then run a fluorescing substance through to see wether there is still a response, if not, clean the cell, if that doesn´t do any good go for the lamp. Still a problem? Call the vendor.
Thanks for the info. I'm trying to clean the detector now with water, and it seems the problem is with air trapped inside the detector. I"ll keep your suggestions for future reference , too.
3 posts Page 1 of 1

Who is online

In total there are 14 users online :: 1 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Baidu [Spider] and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry