no fragmentation in iontrap

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
hi,
i am working on the analysis of oxazepam and oxazepam-glucuronide with lc-msms on an iontrap.
when i bring oxazepam in it gives a nice m+1 in ms and i can fragment this to m+1-H2O.
when i bring oxazepam-glucuronide in it gives a nice m+1 but when i try to fragment it i see no fragments. even with the lowest energy (0%) a see no signal, even the M+1 is gone.
i have seen this before even with compounds i read about that other labs can do MSMS on with an triple-quad.
my questions are
has anybody seen this?
is this an iontrap-thing?
is there anyway i can tweak settings to give a MS/MS-spectrum?

i ame working on an LCQ-classic.

thanks i.a. for your input.

If the instrument will let you, try an isolation-only experiment (no fragmentation cycle). If you still don't see it, then it is being lost in the isolation step. Either way (either with isolation-only or with frag amplitude set to zero) try to increase your isolation window and see if it comes back. I remember one or two compounds that needed a wider isolation window than default, otherwise they would be lost in the isolation step. I couldn't explain the behavior, though.

whow! that did the trick! it works.
but i can not come up whit a good expalanation.
anybody has any idea why this works?
3 posts Page 1 of 1

Who is online

In total there are 39 users online :: 1 registered, 0 hidden and 38 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 38 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry