polystyrene divinylbenzene sulfate columns

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Just wondering if anyone has messed around with using ammonium (NH4+ or other amines) modifiers to get any selectivity change with carbohydrate analyses. I have seen the various salt forms from H+ to Ag+, so just wondering if it is possible to take an H+ form and treat it with an ammonium salt in the mobile phase.
:?: :?: PSDVB sulfonated columns for sugars usually use sodium, calcium or lead to form chelates with sugars to obtain separation. The main eluent for these columns is water. Using acid or base will poison the packing material and destroy separation.

It's virtually impossible to change ionic form of those columns in-situ because the resins shrink/swell. As Dave pointed out, trying to do the conversion (e.g. from the H+ form to the ammonium form) will effectively kill the column.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 20 users online :: 1 registered, 0 hidden and 19 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 19 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry