CTC Pal headspace problem

Discussions about GC and other "gas phase" separation techniques.

2 posts Page 1 of 1
Hi all,

i currently use an Agilent 7890 GC coupled with a 5975MSD and a CTC pal autosampler. i need to develop a method where i use a different vial size (i.e height).

Could somebody tell me how to adjust this setting please?

Many Thanks

Steve

wrighty,

I don't have a CTC but a MPS (CTC modified and sold by Gerstel). They are relatively similar.
You have up load the objects (i.e. sample trays, syringe type, vial type, etc..) with a software called PALoader that should be present somewhere on your pc. Once you have done that you can start to modify the settings using the remote attached to your CTC.

Regards,
bhuvfe