I am operating a Thermo Ultra Trace GC-MS. I recently installed a new column (Rtx-5MS 30mx0.25mmx0.25um). At the same time, I also changed the septum, seal, and ferrules. I performed a column evaluation and the K value was a little low, but when I ran multiple leak checks it always came back as 'passed'.
However, I had the split flow set at 25 ml/min with the carrier gas (helium) set to constant flow of 1 ml/min. I noticed on the GC control panel and on computer screen that the setpoint was set to 25 ml/min while the actual said it was at 180 ml/min (with total flow rate of 185.6 ml/min) with a pressure of about 10 psi.
I have tried turning the GC off and then restarting. I have tried changing the split to different flow rates (50, 100, 150, 200, 250 ml/min) and nothing would change it from 180 ml/min. When I did turn this carrier flow to 'off', the split flow did go to 0 ml/min, so this is a good sign. I tried changing from 'constant flow' to 'constant pressure'. I tried turning the Helium pressure at the gas tank down from 90 psi to 30 psi. None of these worked.
Any suggestions to troubleshoot this would be most greatly appreciated.