I tried this yesterday: here are my retention times for one peak:Doesn't the method require the sample solvent to be 80:20 H2O:CH3CN?. I'd also try dissolving the sample in the mobile phase itself.
inj H2O/ACNMobile phase
1 16.002 15.578
2 15.917 15.829
3 15.775 15.507
4 15.673 15.966
5 15.876 15.862
6 15.405 15.532
Remark: my peaks for my samples dissolved in mobile phase were doubling at the end of my sample set (from injection 5). Immediately after these samples, I began to inject the samples dissolved in H2O/ACN. The retention time increased immediately from 15.532 to 16.00 and my peak shape is back to normal.
After seeing above numbers, its hard to see if the variations change I think.My suspicion is that the repeated injections of samples is affecting retention.
I'd try injecting varying numbers of mobile phase blanks between samples and seeing if the retention time variations change.
I will send the EP experts an email. I haven't tried the Symmetry yet, but after trying 3 different columns (ABZ+, BEH Shield RP18 and a BEH C18), I think we can say that it isn't a column problem?I'd also email the EP experts, and ask them for suggestions. The database also suggests a Waters Symmetry column, and I'd try that.