I am running an HP5890 with the 7673 autosampler, and I am experiencing several issues with my system. The runs are being controled by chemstation revision A.06.01.
Issue 1, regardless of manual or automatic injection the chemstation window never displays that the system is ready. This prompted, obviously, double checking my entire method; all parameters and hardware appear to be sorted correctly (oven equilibrated, injectors and detectors at temp, etc.). When I manually inject a sample and hit start on the GC the run starts straight away and it ignores the "waiting for ready" alert. However, if I set up an autoinjector sequence the sequence will run, but with a catch: I must select the option to ignore the "waiting for ready" alert before the sequence will run and in between subsequent injections. Rendering the autosampler impossible to use without someone there to help it along. I'm unsure what it could be.
Any experiences with this? Also, the GC is hooked up to an integrator; could that alter the software/hardware communications?
Issue 2, again with the autosampler. I've used our autosampler on both splitless and split injection, this requires the ability to change the amount of sample delivered by the autosampler to the column. When I enter the sequence parameters and change the injection volume there is no change to the physical volume delivered. This is still the case after compensating for viscosity. If you are curious the volume that our autosampler chooses is 0.8ul. This volume is okay for the vast majority of samples run with the autoinjector, but see Issue 1.
Could this be an issue of worn our hardware? The equipment (and software installation) pre dates my employment. I thought it could also boil down to poor alignment of the autosampler. Finally, could reinstalling my Chemstation make a difference?
Thanks for any help, best wishes.