Passivation of injection needle

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Do you perform passivation of injection needle when you passivate the rest of the LC?

is it possible to inject 30% HNO3, or will it damage the seals?

depends on the material of the seal but I would say they can handle 30%nitric acid. I have passivated needles and injectors in the past on following instruments:
HP1090, Agilent 1100, Hitachi/Merck La Chrom. In at least one case I had a stubborn contamination. I removed the injection needle from the system and cleaned it in 69%nitric acid for about 30 minute using an ultrasonic bath. Worked really well.

Make sure that the parts are free from any organic solvent (especially alcohols) and use a separated waste container to catch the nitric acid.

Robert
--
Robert Haefele
2 posts Page 1 of 1

Who is online

In total there are 116 users online :: 2 registered, 0 hidden and 114 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Majestic-12 [Bot], Semrush [Bot] and 114 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry