I am currently running Agilent 6890 GC with Chemstation B01.01 (I think, not at work right now), and am having trouble getting the post-sequence commands to work. I am trying to get the shutdown method to load at the end of the sequence. The commands work when entered in the command line, but when placed in the post-sequence cmd/macro line in either the run-time checklist or in the sequence paramenters, it doesn't function.

Has anyone encountered this problem or have any ideas of why the commands don't work as a post-sequence command?

The command I am using is: loadmethod _methpath$, "shutdown.m"

Hi spill,
add one plus line to the end of the sequence, leave the vial# empty or inject a solvent, with the 'shutdown.m' method.
goood luck.
ATH
2 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot], Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry