I'm running a 2.1x100mm C18 column, 2.1x10mm guard column, flow 0.2 ml/min, 60-40 A:B to 0:100 A:B in ten minutes, flush 5 minutes at 100 B, back to starting conditions long enough to equilibrate, A:HPLC grade (bottled) water, B:ACN. Currently have about 700 injections on column, detector is Qtof, ESI-, MS-MS w/fragmentation to quantify. This is the setup we have been running for the last month, samples are groundwater cleaned up via SPE and in 50:50 H2O:ACN. Guard colum was changed about 300 injections ago.
Last week (and for several weeks prior) back pressure has been ~1200 PSI, areas of standards and IS have been fairly consistent (~20 for IS). The last few days, the back pressure has climbed to as high as 1500 and yesterday it was only 1400. Delta PSI is about 15. The areas of our standards and IS have increased 3x in area from yesterday to today. My question is: as a result of "stuff" getting on the column, is the compound binding better, and more is getting held and then released? It seems so drastic for it increase 3x better sensitivity (IS = 60 area units). Retention time is the same for both days and hasn't changed when PSI was 1200 vs the now 1500. I'll cross post in Hyphenated techniques, but it seems like an LC issue more than MS.